Calibration of Rosemount Transmitter
Dec 02, 2021
1. Zero Calibration
When adjusting the transmitter, please turn off the protection setting, which is usually the red switch in front of the meter and its status may be ON or OFF.
If it is ON, you will not be able to make the calibration adjustment.
Rosemount Differential Pressure Transmitter 1151DP, Type Supply:45 VDC MAX Output:4-20mA
Figure 1: Rosemount Transmitter 1151DP5E22M6Y1
Calibration Steps
First, ventilate the transmitter, and connect measuring circuit with the manual operator.
Then, operate on the manual operator as follows.
Calibration Steps
Finally, calibrate the transmitter through zero calibration according to the manual operator.
It is important that the zero calibration must be for the transmitter whose output is within 3% around zero.
2. DAC Calibration (Output Calibration)
The output calibration means to calibrate the signal output conversion from digital-to-analog (DAC).
Operate on the manual operator as follows.
DAC Calibration
And set the control loop as manual mode.
Before confirming, notice that if the positive and negative wiring is correct.
Finally, the output of the field device is set to 4mA, and the output of the transmitter is 4mA. Then record the actual value of the reference table, and input the value on the transmitter.
Check whether the output of the manual operator is consistent with the value of the reference table. If it is consistent, change the output to 20mA, and then repeat the operation until it's accurate. If not, you should continue to input the reference value and inspect whether it is consistent with the manual operator output.
Finally, set the control loop as automatic control.
Rosemount High Static Pressure Transmitter 1151HP Type Supply:45 VDC MAX Output:4-20mA
Figure 2: Rosemount Transmitter 1151HP7S22M4B1
3. ADC Calibration (Sensor Calibration)
The sensor calibration is to calibrate the analog-to-digital converter (ADC). The main work to be done is zero calibration and full calibration. Full calibration requires three times the accuracy of the transmitter. Operate on the manual operator as follows.
ADC Calibration
If the accuracy value of transmitter is 0.25mA, the lower limit output will be 0.06mA.
Related Info
Benefits of Rosemount Fieldbus Transmitter
Rosemount Transmitter Series 1151, 2088, 2051, 3051CA, 3051S
Thank you for you reading and more information please visit https://okmarts.com/news/transmitter.html
Dec 30, 2021 -
Dec 30, 2021 -
Dec 29, 2021 -
Dec 29, 2021 -
Dec 28, 2021 -
Dec 28, 2021 -
Dec 27, 2021 -
Dec 27, 2021 -
Dec 26, 2021 -
Dec 26, 2021 -
Calibration of Rosemount Transmitter news release has been viewed 113 times
Fluid Dispensers